What is IEC 61000-4-4?

What is IEC 61000-4-4?

1 Answer
Can you answer this question?

Editorial Team - EMC Directory

Jul 19, 2022

The IEC 61000-4-4 is an EMC standard by the International Electrotechnical Commission for the immunity requirements of electrical and electronic equipment to repetitive electrical fast transients. This standard specifies the limits, requirements, and procedures to evaluate the performance of the Equipment Under Test (EUT) when exposed to repetitive fast transients (bursts) on supply, signal, and control ports. These transient bursts are the disturbances created by switching power currents, the opening of circuit contacts leading to sparks, and the switching of loads in a circuit. 

IEC 61000-4-4 defines the EUT’s test voltage waveform, range of test levels, test equipment, test set-up, and test procedures. The test equipment to be used according to the standard is:

  • Test Generator
  • Ground reference plane
  • Capacitive coupling clamp
  • Decoupling network

According to the IEC 61000-4-4 standard, EFT test pulses are single 5 kHz unidirectional pulses lasting for 15 milliseconds each, repeated at three bursts per second for about one minute. These pulses are injected into the AC and DC power ports of the device under test using coupling-decoupling networks, whereas capacitive clamps are used for injection into the signal and control ports.

The standard also classifies the test levels for EFT/B testing according to the requirements of the electromagnetic environment. These levels are: 

  1. Level 1: Well-protected Environment e.g., a computer room.
  2. Level 2: Protected Environment e.g., a control room of an industrial plant.
  3. Level 3: Typical industrial Environment e.g., a relay room of open-air, high voltage power plant or substation of manufacturing plant.
  4. Level 4: Severe Industrial Environment e.g., substation switchyards.

The range of test levels for the electrical fast transient test, applicable to power supply, protective earth (PE), signal and control ports of the equipment as per IEC 61000-4-4 are given as follows. The test levels are measured on the open-circuit output test voltage (ą10 %) and the repetition rate of the impulses (ą20 %).


LevelOn I/O signal, data, and control portsRepetition Rate

Peak Voltage (kV)(kHz)
10.50.255 or 100
210.55 or 100
3215 or 100
4425 or 100

Table: Test levels for EFT Test

Click here for IEC 61000-4-4 Testing Labs.

Technical Articles

../ 26 [{"category":"Questions","label":"what is cispr 16?","url":"../community/what-is-cispr-16","type":null},{"category":"Questions","label":"what is fcc part 15?","url":"../community/what-is-fcc-part-15","type":null},{"category":"Questions","label":"what is iec 61000-6-4?","url":"../community/what-is-iec-61000-6-4","type":null},{"category":"Questions","label":"what is fcc part 18?","url":"../community/what-is-fcc-part-18","type":null},{"category":"Questions","label":"what is iec 61000-4-3?","url":"../community/what-is-iec-61000-4-3","type":null},{"category":"Questions","label":"what is iec 61000-4-8?","url":"../community/what-is-iec-61000-4-8","type":null},{"category":"Questions","label":"what is iec 61000-4-4?","url":"../community/what-is-iec-61000-4-4","type":null},{"category":"Questions","label":"what is iec 61000-4-6?","url":"../community/what-is-iec-61000-4-6","type":null},{"category":"Questions","label":"what is cispr 32?","url":"../community/what-is-cispr-32","type":null},{"category":"Questions","label":"what is iec 61000-4-11?","url":"../community/what-is-iec-61000-4-11","type":null},{"category":"Questions","label":"what is iec 61000-4-2?","url":"../community/what-is-iec-61000-4-2","type":null},{"category":"Questions","label":"what is iec 61000-4-5?","url":"../community/what-is-iec-61000-4-5","type":null},{"category":"Questions","label":"what is radiated emissions testing?","url":"../community/what-is-radiated-emission-testing","type":null},{"category":"Questions","label":"what is iec 61000-3-3?","url":"../community/what-is-iec-61000-3-3","type":null},{"category":"Questions","label":"what is conducted emissions testing?","url":"../community/what-is-conducted-emissions-testing","type":null},{"category":"Questions","label":"what is radiated immunity testing?","url":"../community/what-is-radiated-immunity-testing","type":null},{"category":"Questions","label":"what is cispr 22?","url":"../community/what-is-cispr-22","type":null},{"category":"Questions","label":"what is voltage drop testing?","url":"../community/what-is-voltage-drop-testing","type":null},{"category":"Questions","label":"what are electro-static discharge(esd) immunity tests?","url":"../community/what-is-esd-testing","type":null},{"category":"Questions","label":"what is magnetic field testing?","url":"../community/what-is-magnetic-field-testing","type":null},{"category":"Questions","label":"what is open area test site (oats)?","url":"../community/what-is-open-area-test-site-oats","type":null},{"category":"Questions","label":"what is harmonic and flicker testing?","url":"../community/what-is-harmonic-and-flicker-testing","type":null},{"category":"Questions","label":"what is iec 61000-4-9?","url":"../community/what-is-iec-61000-4-9","type":null},{"category":"Questions","label":"what is the difference between emc and emi?","url":"../community/what-is-the-difference-between-emc-and-emi","type":null},{"category":"Questions","label":"what is emc testing? why is it important?","url":"../community/what-is-emc-testing-why-is-it-important","type":null},{"category":"Questions","label":"emc testing for household appliances","url":"../community/what-is-emc-testing-for-household-appliances","type":null},{"category":"Questions","label":"what is surge immunity testing?","url":"../community/what-is-surge-immunity-testing","type":null},{"category":"Questions","label":"what is eft immunity testing?","url":"../community/what-is-eft-immunity-testing","type":null},{"category":"Questions","label":"what is iec 61000-3-2?","url":"../community/what-is-iec-61000-3-2","type":null},{"category":"Questions","label":"what is cispr 12?","url":"../community/what-is-cispr-12_12","type":null},{"category":"Questions","label":"what is cispr 25?","url":"../community/what-is-cispr-25_13","type":null},{"category":"Questions","label":"what is cispr 35?","url":"../community/what-is-cispr-35_14","type":null},{"category":"Questions","label":"what are emc emissions tests?","url":"../community/what-is-emc-emissions-tests","type":null},{"category":"Questions","label":"what is emc pre-compliance testing?","url":"../community/what-is-emc-pre-compliance-testing","type":null},{"category":"Questions","label":"what is immunity testing for emc?","url":"../community/what-is-immunity-testing-for-emc","type":null},{"category":"Questions","label":"what is cispr 31?","url":"../community/what-is-cispr-31","type":null},{"category":"Questions","label":"emc testing for a medical device","url":"../community/what-is-emc-testing-for-a-medical-device","type":null},{"category":"Questions","label":"what are helmholtz coils?","url":"../community/what-are-helmholtz-coils","type":null},{"category":"Questions","label":"emc testing for automotive","url":"../community/what-is-emc-testing-for-automotive","type":null},{"category":"Questions","label":"what is conductive coupling?","url":"../community/what-is-conductive-coupling","type":null},{"category":"Questions","label":"what is inductive coupling?","url":"../community/what-is-inductive-coupling","type":null},{"category":"Questions","label":"what is a tem cell?","url":"../community/what-is-a-tem-cell","type":null},{"category":"Questions","label":"emc testing for printed circuit boards (pcbs)","url":"../community/what-is-emc-testing-for-pcb-what-are-the-design-consideration","type":null},{"category":"Questions","label":"emc testing for iot/wireless devices","url":"../community/what-is-emc-testing-for-iot-wireless-devices","type":null},{"category":"Questions","label":"what is an emc test site?","url":"../community/what-is-an-emc-test-site","type":null},{"category":"Questions","label":"iso 11452: emc testing standards for automotive components","url":"../community/what-is-iso-11452","type":null},{"category":"Questions","label":"what is crosstalk?","url":"../community/what-is-crosstalk","type":null},{"category":"Questions","label":"what is near-field shielding?","url":"../community/what-is-near-field-shielding","type":null},{"category":"Questions","label":"what is far-field shielding?","url":"../community/what-is-far-field-shielding","type":null},{"category":"Questions","label":"what is cispr 11?","url":"../community/what-is-cispr-11","type":null},{"category":"Questions","label":"what is iec 61000-6-2?","url":"../community/what-is-iec-61000-6-2","type":null},{"category":"Questions","label":"what is iec 61000-6-3?","url":"../community/what-is-iec-61000-6-3","type":null},{"category":"Questions","label":"what is pulse width modulation (pwm)?","url":"../community/what-is-pulse-width-modulation-pwm","type":null},{"category":"Questions","label":"what are common mode chokes?","url":"../community/what-are-common-mode-chokes","type":null},{"category":"Questions","label":"what is iso/iec 17025?","url":"../community/what-is-iso-17025","type":null},{"category":"Questions","label":"what is esd?","url":"../community/what-is-esd","type":null},{"category":"Questions","label":"what is mil-std-461?","url":"../community/what-is-mil-std-461","type":null},{"category":"Questions","label":"what is rtca do-160?","url":"../community/what-is-rtca-do-160","type":null},{"category":"Questions","label":"what is def-stan 59-411?","url":"../community/what-is-def-stan-59-411","type":null},{"category":"Questions","label":"what is iso 14302?","url":"../community/what-is-iso-14302","type":null},{"category":"Questions","label":"fast measurements with real-time spectrum analyzers for emc testing","url":"../community/fast-measurements-with-real-time-spectrum-analyzers-for-emc-testing","type":null},{"category":"Questions","label":"what are semi-anechoic chambers?","url":"../community/what-are-semi-anechoic-chambers","type":null},{"category":"Questions","label":"what is electromagnetic interference (emi)?","url":"../community/what-is-electromagnetic-interference-emi","type":null},{"category":"Questions","label":"what is radio frequency interference (rfi)?","url":"../community/what-is-radio-frequency-interference-rfi","type":null},{"category":"Questions","label":"what is iec 61967-4?","url":"../community/what-is-iec-61967-4","type":null},{"category":"Questions","label":"what is iec 61967-5?","url":"../community/what-is-iec-61967-5","type":null},{"category":"Questions","label":"what are the emc testing standards specification for integrated circuits?","url":"../community/what-are-the-standardized-emc-test-methods-standards-published-by-iec-for-integrated-circuits-ics","type":null},{"category":"Questions","label":"what is iec 61000-6-1?","url":"../community/what-is-iec-61000-6-1","type":null},{"category":"Questions","label":"what is iec 61000-6-2?","url":"../community/what-is-iec-61000-6-2_102","type":null},{"category":"Questions","label":"what is iec 61000-4-39?","url":"../community/what-is-iec-61000-4-39","type":null},{"category":"Questions","label":"what is characteristic impedance?","url":"../community/what-is-characteristic-impedance","type":null},{"category":"Questions","label":"what is common-mode noise?","url":"../community/what-is-common-mode-noise","type":null},{"category":"Questions","label":"what is a ferrite clamp?","url":"../community/what-is-the-ferrite-clamp","type":null},{"category":"Questions","label":"what is conducted immunity testing?","url":"../community/what-is-conducted-immunity-testing","type":null},{"category":"Questions","label":"what is iec 61967-1?","url":"../community/what-is-iec-61967-1","type":null},{"category":"Questions","label":"what is iec 61967-2?","url":"../community/what-is-iec-61967-2","type":null},{"category":"Questions","label":"what are the differences between anechoic and semi-anechoic chambers?","url":"../community/what-are-the-differences-between-anechoic-and-semi-anechoic-chambers","type":null},{"category":"Questions","label":"what are anechoic chambers?","url":"../community/what-are-anechoic-chambers","type":null},{"category":"Questions","label":"what is mil-std-1275?","url":"../community/what-is-mil-std-1275","type":null},{"category":"Questions","label":"what is iec 60601-1-2?","url":"../community/what-is-iec-60601-1-2","type":null},{"category":"Questions","label":"what is ansi c63.10?","url":"../community/what-is-ansi-c63-10","type":null},{"category":"Questions","label":"what is ansi c63.4?","url":"../community/what-is-ansi-c63-4","type":null},{"category":"Questions","label":"what is an emi filter?","url":"../community/what-is-an-emi-filter","type":null},{"category":"Questions","label":"what is an electromagnetic reverberation chamber?","url":"../community/what-is-an-electromagnetic-reverberation-chamber","type":null},{"category":"Questions","label":"what are coupling decoupling networks in emc?","url":"../community/what-are-coupling-decoupling-networks-in-emc","type":null},{"category":"Questions","label":"what is cispr 14?","url":"../community/what-is-cispr-14","type":null},{"category":"Questions","label":"what is cipsr 36?","url":"../community/what-is-cipsr-36","type":null},{"category":"Questions","label":"shielding and grounding techniques for reducing emi and ems","url":"../community/what-are-the-shielding-and-grounding-techniques-for-reducing-emi-and-ems","type":null},{"category":"Questions","label":"what is cispr 17?","url":"../community/what-is-cispr-17","type":null},{"category":"Questions","label":"what is cipsr 15?","url":"../community/what-is-cipsr-15","type":null},{"category":"Questions","label":"what is iec 61000-6-8?","url":"../community/what-is-iec-61000-6-8","type":null},{"category":"Questions","label":"what is iec 61000-6-4?","url":"../community/what-is-iec-61000-6-4_57","type":null},{"category":"Questions","label":"what is iec 61000-6-3?","url":"../community/what-is-iec-61000-6-3_58","type":null},{"category":"Questions","label":"what is conducted susceptibility?","url":"../community/what-is-conducted-susceptibility","type":null},{"category":"Questions","label":"what is cispr 18?","url":"../community/what-is-cispr-18","type":null},{"category":"Questions","label":"what is power line transient?","url":"../community/what-is-a-power-line-transient","type":null},{"category":"Questions","label":"what are emc antennas?","url":"../community/what-are-emc-antennas","type":null},{"category":"Questions","label":"what is iec 61967-3?","url":"../community/what-is-iec-61967-3","type":null},{"category":"Questions","label":"what is iec 61967-6?","url":"../community/what-is-iec-61967-6","type":null},{"category":"Questions","label":"what is iec 61967-8?","url":"../community/what-is-iec-61967-8","type":null},{"category":"Questions","label":"what is the directive 2014/30/eu?","url":"../community/what-is-the-directive-2014-30-eu","type":null},{"category":"Questions","label":"what is ieee/ansi c63.15?","url":"../community/what-is-ieee-ansi-c63-15","type":null},{"category":"Questions","label":"what is propagation constant?","url":"../community/what-is-propagation-constant","type":null},{"category":"Questions","label":"what is a transmission line? balanced vs unbalanced transmission lines?","url":"../community/what-is-a-transmission-line-balanced-vs-unbalanced-transmission-lines","type":null},{"category":"Questions","label":"what is ieee c95.3?","url":"../community/what-is-ieee-c95-3","type":null},{"category":"Questions","label":"what is fcc mp-5-1986?","url":"../community/what-is-fcc-mp-5-1986","type":null},{"category":"Questions","label":"what is ieee/ansi c63.17?","url":"../community/what-is-ieee-ansi-c63-17","type":null},{"category":"Questions","label":"what is radiated emission?","url":"../community/what-is-radiated-emission","type":null},{"category":"Questions","label":"what is electromagnetic pulse (emp)?","url":"../community/what-is-electromagnetic-pulse-emp","type":null},{"category":"Topics","label":"cispr standards","url":"~/community/filter?topic=;CISPR Standards;","type":null},{"category":"Topics","label":"conducted immunity","url":"~/community/filter?topic=;Conducted Immunity;","type":null},{"category":"Topics","label":"emission testing","url":"~/community/filter?topic=;Emission Testing;","type":null},{"category":"Topics","label":"iec standards","url":"~/community/filter?topic=;IEC Standards;","type":null},{"category":"Topics","label":"immunity testing","url":"~/community/filter?topic=;Immunity Testing;","type":null},{"category":"Topics","label":"medical","url":"~/community/filter?topic=;Medical;","type":null},{"category":"Topics","label":"military","url":"~/community/filter?topic=;Military;","type":null},{"category":"Topics","label":"radiated immunity","url":"~/community/filter?topic=;Radiated Immunity;","type":null}] community community question-details content_more