What is IEC 61000-4-2?

What is IEC 61000-4-2? What is its scope? What are the test levels required?

1 Answer
Can you answer this question?

Editorial Team - EMC Directory

Dec 19, 2022

The IEC 61000-4-2 is an electromagnetic standard by the International Electrotechnical Commission that specifies the limits, requirements, and test methods for electrical and electronic equipment subjected to static electricity discharges, from operators directly, and from personnel to adjacent objects. All people carry static electromagnetic discharge that gives rise to transient currents and electromagnetic fields that can interfere with electrical and electronic equipment. Hence, this standard for established to ensure that the equipment will work reliably and efficiently despite the static interferences in their work environment.

The standard has defined four levels of ESD protection, using two different testing methods. The first is the contact discharge and the second is air discharge. The contact discharge involves sending an ESD pulse directly from the ESD test gun that is touching the equipment under test (EUT). This is the preferred method of testing. Air discharge is used for cases where contact discharge testing is not possible. In the air discharge test, the ESD test gun is brought close to the EUT until a discharge occurs. Both methods have respective limits which are similar to each other for every level.

Contact Discharge
Air Discharge
Level
Test Voltage (kV)
Level
Test Voltage (kV)
1
2
1
2
2
4
2
4
3
6
3
8
4
8
4
15

The first edition of this standard was released in 1995. It was updated with multiple amendments until a new edition 2.0 was released in 2008. The 2.0 edition is still valid. IEC 61000-4-2:2008 has defined a typical waveform of the discharge current, range of test levels, test equipment, test setup, test procedure, calibration procedure, and measurement uncertainty. The latest edition has given specifications for the test performed in laboratories and also for the post-installation tests which are needed to be performed on equipment in the final installation. The second edition has brought equipment up to 4 GHz of frequency under it. An example of EUT matching the requirements is also provided in the 2nd edition. It has provided requirements and procedures on radiated fields from human-metal discharge and ESD generators.

Click here for IEC 61000-4-2 test labs.

Technical Articles

../ 17 [{"category":"Questions","label":"what is cispr 16?","url":"../community/what-is-cispr-16","type":null},{"category":"Questions","label":"what is fcc part 15?","url":"../community/what-is-fcc-part-15","type":null},{"category":"Questions","label":"what is iec 61000-6-4?","url":"../community/what-is-iec-61000-6-4","type":null},{"category":"Questions","label":"what is fcc part 18?","url":"../community/what-is-fcc-part-18","type":null},{"category":"Questions","label":"what is iec 61000-4-3?","url":"../community/what-is-iec-61000-4-3","type":null},{"category":"Questions","label":"what is iec 61000-4-8?","url":"../community/what-is-iec-61000-4-8","type":null},{"category":"Questions","label":"what is iec 61000-4-4?","url":"../community/what-is-iec-61000-4-4","type":null},{"category":"Questions","label":"what is iec 61000-4-6?","url":"../community/what-is-iec-61000-4-6","type":null},{"category":"Questions","label":"what is cispr 32?","url":"../community/what-is-cispr-32","type":null},{"category":"Questions","label":"what is iec 61000-4-11?","url":"../community/what-is-iec-61000-4-11","type":null},{"category":"Questions","label":"what is iec 61000-4-2?","url":"../community/what-is-iec-61000-4-2","type":null},{"category":"Questions","label":"what is iec 61000-4-5?","url":"../community/what-is-iec-61000-4-5","type":null},{"category":"Questions","label":"what is radiated emissions testing?","url":"../community/what-is-radiated-emission-testing","type":null},{"category":"Questions","label":"what is iec 61000-3-3?","url":"../community/what-is-iec-61000-3-3","type":null},{"category":"Questions","label":"what is conducted emissions testing?","url":"../community/what-is-conducted-emissions-testing","type":null},{"category":"Questions","label":"what is radiated immunity testing?","url":"../community/what-is-radiated-immunity-testing","type":null},{"category":"Questions","label":"what is cispr 22?","url":"../community/what-is-cispr-22","type":null},{"category":"Questions","label":"what is voltage drop testing?","url":"../community/what-is-voltage-drop-testing","type":null},{"category":"Questions","label":"what are electro-static discharge(esd) immunity tests?","url":"../community/what-is-esd-testing","type":null},{"category":"Questions","label":"what is magnetic field testing?","url":"../community/what-is-magnetic-field-testing","type":null},{"category":"Questions","label":"what is open area test site (oats)?","url":"../community/what-is-open-area-test-site-oats","type":null},{"category":"Questions","label":"what is harmonic and flicker testing?","url":"../community/what-is-harmonic-and-flicker-testing","type":null},{"category":"Questions","label":"what is iec 61000-4-9?","url":"../community/what-is-iec-61000-4-9","type":null},{"category":"Questions","label":"what is the difference between emc and emi?","url":"../community/what-is-the-difference-between-emc-and-emi","type":null},{"category":"Questions","label":"what is emc testing? why is it important?","url":"../community/what-is-emc-testing-why-is-it-important","type":null},{"category":"Questions","label":"emc testing for household appliances","url":"../community/what-is-emc-testing-for-household-appliances","type":null},{"category":"Questions","label":"what is surge immunity testing?","url":"../community/what-is-surge-immunity-testing","type":null},{"category":"Questions","label":"what is eft immunity testing?","url":"../community/what-is-eft-immunity-testing","type":null},{"category":"Questions","label":"what is iec 61000-3-2?","url":"../community/what-is-iec-61000-3-2","type":null},{"category":"Questions","label":"what is cispr 12?","url":"../community/what-is-cispr-12_12","type":null},{"category":"Questions","label":"what is cispr 25?","url":"../community/what-is-cispr-25_13","type":null},{"category":"Questions","label":"what is cispr 35?","url":"../community/what-is-cispr-35_14","type":null},{"category":"Questions","label":"what are emc emissions tests?","url":"../community/what-is-emc-emissions-tests","type":null},{"category":"Questions","label":"what is emc pre-compliance testing?","url":"../community/what-is-emc-pre-compliance-testing","type":null},{"category":"Questions","label":"what is immunity testing for emc?","url":"../community/what-is-immunity-testing-for-emc","type":null},{"category":"Questions","label":"what is cispr 31?","url":"../community/what-is-cispr-31","type":null},{"category":"Questions","label":"emc testing for a medical device","url":"../community/what-is-emc-testing-for-a-medical-device","type":null},{"category":"Questions","label":"what are helmholtz coils?","url":"../community/what-are-helmholtz-coils","type":null},{"category":"Questions","label":"emc testing for automotive","url":"../community/what-is-emc-testing-for-automotive","type":null},{"category":"Questions","label":"what is conductive coupling?","url":"../community/what-is-conductive-coupling","type":null},{"category":"Questions","label":"what is inductive coupling?","url":"../community/what-is-inductive-coupling","type":null},{"category":"Questions","label":"what is a tem cell?","url":"../community/what-is-a-tem-cell","type":null},{"category":"Questions","label":"emc testing for printed circuit boards (pcbs)","url":"../community/what-is-emc-testing-for-pcb-what-are-the-design-consideration","type":null},{"category":"Questions","label":"emc testing for iot/wireless devices","url":"../community/what-is-emc-testing-for-iot-wireless-devices","type":null},{"category":"Questions","label":"what is an emc test site?","url":"../community/what-is-an-emc-test-site","type":null},{"category":"Questions","label":"iso 11452: emc testing standards for automotive components","url":"../community/what-is-iso-11452","type":null},{"category":"Questions","label":"what is crosstalk?","url":"../community/what-is-crosstalk","type":null},{"category":"Questions","label":"what is near-field shielding?","url":"../community/what-is-near-field-shielding","type":null},{"category":"Questions","label":"what is far-field shielding?","url":"../community/what-is-far-field-shielding","type":null},{"category":"Questions","label":"what is cispr 11?","url":"../community/what-is-cispr-11","type":null},{"category":"Questions","label":"what is iec 61000-6-2?","url":"../community/what-is-iec-61000-6-2","type":null},{"category":"Questions","label":"what is iec 61000-6-3?","url":"../community/what-is-iec-61000-6-3","type":null},{"category":"Questions","label":"what is pulse width modulation (pwm)?","url":"../community/what-is-pulse-width-modulation-pwm","type":null},{"category":"Questions","label":"what are common mode chokes?","url":"../community/what-are-common-mode-chokes","type":null},{"category":"Questions","label":"what is iso/iec 17025?","url":"../community/what-is-iso-17025","type":null},{"category":"Questions","label":"what is esd?","url":"../community/what-is-esd","type":null},{"category":"Questions","label":"what is mil-std-461?","url":"../community/what-is-mil-std-461","type":null},{"category":"Questions","label":"what is rtca do-160?","url":"../community/what-is-rtca-do-160","type":null},{"category":"Questions","label":"what is def-stan 59-411?","url":"../community/what-is-def-stan-59-411","type":null},{"category":"Questions","label":"what is iso 14302?","url":"../community/what-is-iso-14302","type":null},{"category":"Questions","label":"fast measurements with real-time spectrum analyzers for emc testing","url":"../community/fast-measurements-with-real-time-spectrum-analyzers-for-emc-testing","type":null},{"category":"Questions","label":"what are semi-anechoic chambers?","url":"../community/what-are-semi-anechoic-chambers","type":null},{"category":"Questions","label":"what is electromagnetic interference (emi)?","url":"../community/what-is-electromagnetic-interference-emi","type":null},{"category":"Questions","label":"what is radio frequency interference (rfi)?","url":"../community/what-is-radio-frequency-interference-rfi","type":null},{"category":"Questions","label":"what is iec 61967-4?","url":"../community/what-is-iec-61967-4","type":null},{"category":"Questions","label":"what is iec 61967-5?","url":"../community/what-is-iec-61967-5","type":null},{"category":"Questions","label":"what are the emc testing standards specification for integrated circuits?","url":"../community/what-are-the-standardized-emc-test-methods-standards-published-by-iec-for-integrated-circuits-ics","type":null},{"category":"Questions","label":"what is iec 61000-6-1?","url":"../community/what-is-iec-61000-6-1","type":null},{"category":"Questions","label":"what is iec 61000-6-2?","url":"../community/what-is-iec-61000-6-2_102","type":null},{"category":"Questions","label":"what is iec 61000-4-39?","url":"../community/what-is-iec-61000-4-39","type":null},{"category":"Questions","label":"what is characteristic impedance?","url":"../community/what-is-characteristic-impedance","type":null},{"category":"Questions","label":"what is common-mode noise?","url":"../community/what-is-common-mode-noise","type":null},{"category":"Questions","label":"what is a ferrite clamp?","url":"../community/what-is-the-ferrite-clamp","type":null},{"category":"Questions","label":"what is conducted immunity testing?","url":"../community/what-is-conducted-immunity-testing","type":null},{"category":"Questions","label":"what is iec 61967-1?","url":"../community/what-is-iec-61967-1","type":null},{"category":"Questions","label":"what is iec 61967-2?","url":"../community/what-is-iec-61967-2","type":null},{"category":"Questions","label":"what are the differences between anechoic and semi-anechoic chambers?","url":"../community/what-are-the-differences-between-anechoic-and-semi-anechoic-chambers","type":null},{"category":"Questions","label":"what are anechoic chambers?","url":"../community/what-are-anechoic-chambers","type":null},{"category":"Questions","label":"what is mil-std-1275?","url":"../community/what-is-mil-std-1275","type":null},{"category":"Questions","label":"what is iec 60601-1-2?","url":"../community/what-is-iec-60601-1-2","type":null},{"category":"Questions","label":"what is ansi c63.10?","url":"../community/what-is-ansi-c63-10","type":null},{"category":"Questions","label":"what is ansi c63.4?","url":"../community/what-is-ansi-c63-4","type":null},{"category":"Questions","label":"what is an emi filter?","url":"../community/what-is-an-emi-filter","type":null},{"category":"Questions","label":"what is an electromagnetic reverberation chamber?","url":"../community/what-is-an-electromagnetic-reverberation-chamber","type":null},{"category":"Questions","label":"what are coupling decoupling networks in emc?","url":"../community/what-are-coupling-decoupling-networks-in-emc","type":null},{"category":"Questions","label":"what is cispr 14?","url":"../community/what-is-cispr-14","type":null},{"category":"Questions","label":"what is cipsr 36?","url":"../community/what-is-cipsr-36","type":null},{"category":"Questions","label":"shielding and grounding techniques for reducing emi and ems","url":"../community/what-are-the-shielding-and-grounding-techniques-for-reducing-emi-and-ems","type":null},{"category":"Questions","label":"what is cispr 17?","url":"../community/what-is-cispr-17","type":null},{"category":"Questions","label":"what is cipsr 15?","url":"../community/what-is-cipsr-15","type":null},{"category":"Questions","label":"what is iec 61000-6-8?","url":"../community/what-is-iec-61000-6-8","type":null},{"category":"Questions","label":"what is iec 61000-6-4?","url":"../community/what-is-iec-61000-6-4_57","type":null},{"category":"Questions","label":"what is iec 61000-6-3?","url":"../community/what-is-iec-61000-6-3_58","type":null},{"category":"Questions","label":"what is conducted susceptibility?","url":"../community/what-is-conducted-susceptibility","type":null},{"category":"Questions","label":"what is cispr 18?","url":"../community/what-is-cispr-18","type":null},{"category":"Questions","label":"what is power line transient?","url":"../community/what-is-a-power-line-transient","type":null},{"category":"Questions","label":"what are emc antennas?","url":"../community/what-are-emc-antennas","type":null},{"category":"Questions","label":"what is iec 61967-3?","url":"../community/what-is-iec-61967-3","type":null},{"category":"Questions","label":"what is iec 61967-6?","url":"../community/what-is-iec-61967-6","type":null},{"category":"Questions","label":"what is iec 61967-8?","url":"../community/what-is-iec-61967-8","type":null},{"category":"Questions","label":"what is the directive 2014/30/eu?","url":"../community/what-is-the-directive-2014-30-eu","type":null},{"category":"Questions","label":"what is ieee/ansi c63.15?","url":"../community/what-is-ieee-ansi-c63-15","type":null},{"category":"Questions","label":"what is propagation constant?","url":"../community/what-is-propagation-constant","type":null},{"category":"Questions","label":"what is a transmission line? balanced vs unbalanced transmission lines?","url":"../community/what-is-a-transmission-line-balanced-vs-unbalanced-transmission-lines","type":null},{"category":"Questions","label":"what is ieee c95.3?","url":"../community/what-is-ieee-c95-3","type":null},{"category":"Questions","label":"what is fcc mp-5-1986?","url":"../community/what-is-fcc-mp-5-1986","type":null},{"category":"Questions","label":"what is ieee/ansi c63.17?","url":"../community/what-is-ieee-ansi-c63-17","type":null},{"category":"Questions","label":"what is radiated emission?","url":"../community/what-is-radiated-emission","type":null},{"category":"Questions","label":"what is electromagnetic pulse (emp)?","url":"../community/what-is-electromagnetic-pulse-emp","type":null},{"category":"Topics","label":"cispr standards","url":"~/community/filter?topic=;CISPR Standards;","type":null},{"category":"Topics","label":"conducted immunity","url":"~/community/filter?topic=;Conducted Immunity;","type":null},{"category":"Topics","label":"emission testing","url":"~/community/filter?topic=;Emission Testing;","type":null},{"category":"Topics","label":"iec standards","url":"~/community/filter?topic=;IEC Standards;","type":null},{"category":"Topics","label":"immunity testing","url":"~/community/filter?topic=;Immunity Testing;","type":null},{"category":"Topics","label":"medical","url":"~/community/filter?topic=;Medical;","type":null},{"category":"Topics","label":"military","url":"~/community/filter?topic=;Military;","type":null},{"category":"Topics","label":"radiated immunity","url":"~/community/filter?topic=;Radiated Immunity;","type":null}] community community question-details content_more